Argon as the carrier gas.
I would like to ask for your comments on the argon used as a carrier gas - optimal velocity, used pressures. There is not much to find in Google. I found only the calculator for GC from Agilent, according to him the optimal velocity is about 10 cm/s.
Regards
Regards
You will find pressures there as well :-)
Regards
10 cm/s is pretty slow about 1/3 of helium. Obviously it cannot be used for MSD work.
Why not use nitrogen a bit faster and cheaper than Argon?
Or, if you are OK with it hydrogen gas. Fast, efficient.
Back when I was using Ar in quantity to run my ICP-MS; I also used Ar as the carrier gas for my GC-FID instrument. It worked OK for that.
However, very bad gas to try for GC-MS. Argon quenches and skews the intensities of compound fragmentation. Pure argon for P&T and carrier gas pretty much makes GC-MS unuseable.
We recently received three tanks of Helium contaminated with Argon, yes it is terrible on the GCMS with purge and trap even at ppm levels.