chromquest 5.0
4 posts Page 1 of 1
I bought a gc from Thermofisher and it uses chromquest 5.0. I cannot manually integrate like one does with waters or agilent software. I can't graphically "click and drag" a baseline with chromquest. Am I missing something?
vestel b. shirley, president
betves inc
166 norwood drive
reidsville, nc 27320
Yes contact your Thermo Support contact as you can manually integrate in ChromQuest as it is the same as EzChrom
I want to click and drag ike one can do with waters or agilent software.
I can do the other - go into the menus, manual peak, start, end, analyze.
5 steps - not two.
betves inc
166 norwood drive
Right click on chromatogram > Graphical Programming > Manual Peak
Alternatively, click the 16th icon from left-to-right at the bottom of the chromatogram.
Also, keep in mind that in Chromquest/EZChrom you have all the option that you find in Chemstation or Waters software, and more. You just have to "learn" the software.