Agilent 1100 autosampler
Does anyone know if the screw rods in the Agilent 1100 autosampler which move the transport need greasing or lubrication? I couldn't find anything in the Reference Manual about this. Thanks.
I have a suspicion that they might. I think it could not hurt to use a little lithium grease on the z- and x-axis rods at the back of the transport assembly. I have done this myself without complication on the z-axis rod. I belive this might avoid some of the 'transport motor overheat' errors that can happen, which usually result in Agilent recommending you replace a $1900 transport assembly.
Time flies like an arrow. Fruit flies like a banana.
I was told by an Agilent Service engineer that these rods should not be greased since this will lead to a build up of dust on the rods which will cause errors. He recomended cleaning the rods with a swab soaked in IPA every few months. We have been doing this for several years and have had no problems.
"He recomended cleaning the rods with a swab soaked in IPA every few months. We have been doing this for several years and have had no problems."
That has to be a pretty annoying task. Do you have to pull the transport assembly out of the autosampler to do this? I can't imagine how you would get to the screw rods any other way.
I have also heard the warning about lubrication, but don't have any further details. If you have a dusty lab I would worry about things sticking.
I have used the IPA approach to help with a corrosion problem (just soaked a soft cloth with IPA and wiped the rods as well as I could without removing the assembly). Seems to work fine now.
A wipe with a silicone spray lubricant might be a better compromise.
Merlin K. L. Bicking, Ph.D.
ACCTA, Inc.
ACCTA, Inc.
In total there are 30 users online :: 1 registered, 0 hidden and 29 guests (based on users active over the past 5 minutes)
Users browsing this forum: Semrush [Bot] and 29 guests
