USP Plate Count
Is the a guideline on plate count for system suitability.
CDER recommends >2000
If I have a plate count of the first analyte of 1700, is the system suitable.
Jay
2000 is a recommendation, not a requirement.
If this is a validated method and its system suitability criteria call for 2000 plates, then no, your system is not suitable.
If you are developing and validating a method, then you should put in appropriate system suitability parameters. If it is a simple assay with no interfering peaks, and you can prove that the method meets your criteria for accuracy, precision, etc., then 1700 may be adequate. But, because your plate count is lower than the "recommendation", you should expect some difficulty from reviewers!
In total there are 28 users online :: 1 registered, 0 hidden and 27 guests (based on users active over the past 5 minutes)
Users browsing this forum: Semrush [Bot] and 27 guests