Linearity of impurities!
Hello experts!
I have a doubt, and I hope that you can help me! I am validating a method for impurities, in the accuracy experiment I get a r2= 0.95, the 100% of the concentration of the sample is 0.1 ug/ml. Can I considered this value of r2 as a linear or I have to change the method? :)
Regards
No single/simple answer to that one. r2 does not *prove* linearity, it merely gives an indication of the degree of scatter. Comparing the r2 values from a linear fit to those from a quadratic fit will give you a bit more insight (but not much). There are more meaningful statistical tests that can be applied. Download and read the articles in this series for more detailed information:
http://www.lcresources.com/sandbox/amla ... stics.html
Tom!
Thank you, the link was very usefull. Now I have more tools for a successfull validation.
Regards
In total there are 24 users online :: 2 registered, 0 hidden and 22 guests (based on users active over the past 5 minutes)
Users browsing this forum: Google [Bot], Semrush [Bot] and 22 guests