Selection of internal standard
I contracted out the synthesis of some internal standards for few metabolites of a compound that I am working on.
The company responsible for the synthesis prepared an important intermediate for generating the internal standards I need.
However they received a mixture of different label grades:
M+5: 50.5%,
M+4: 27.9%,
M+3: 8.6%,
M+2: 0.09%,
no M+1 or below was observed.
Compounds contains 4x deuterium + 1x 15N.
Thinking about a mass spec analysis which I will select a specific mass transition for the native compound (unlabelled) and for the internal standard (and that generally I use a high concentration of internal standard compared to the analytical curve), would this different grades be an issue?
I do not think so, as even considering that I only have 50% M+5, I will probably have enough signal in the equipment, and since I'll be selecting an specific m/z I believe no interference will be observed. Is that correct?
I agree with you. As long as there is no interference in both the unlabelled and labelled acqusition functions, and your labelled molecule is stable over time, this should be fine.
Users browsing this forum: Bing [Bot], Semrush [Bot] and 5 guests