Sargent Welch XKR to HP 5890 connetion
How can I connect a Sargent Welch Recorder Model XKR to an HP 5890 GC? I do not have any serial cables that came with the unit. My predecessor used such a connection but that was in 2000 when he passed and all remains that have been misplaced. Any help on what is needed or to look for to connect?
I could have simply told you how to do this- 25 years ago.
If you post your E-mail address, I'll send you the 5890 manual in PDF. You'll need Chapter 6: Controlling signal output. I saw on p. 170 where it showed output of 0 to +0.1 mV for strip chart recorders.
Install a new communication board 19257-60010 or 19242-60030 ( ca be found on ebay)
https://www.ebay.com/sch/i.html?_from=R ... &_osacat=0
And try our UniChrom https://www.unichrom.com/hpx890/hpx890e.shtml
WillieBill,
Did you get this resolved? KM-US is correct, the pin outs for the 5890 are in the manual and would allow you to hook to a strip chart recorder. I would rather recommend an electronic data system and there are several that are inexpensive and allow you to use a computer. Hugely advantageous over a strip chart. Can provide examples if you would like....
The other issue is, the pin outs on the 5890 limit you to just 0-1V and greatly limit your dynamic range. The detector is capable from ppm to part per 10K but the output for that is way beyond 0-1v. Unless you are working in one concentration range you are not taking advantage of the full dynamic range of the detector.
Best regards,
AICMM
Best regards,
AICMM
Users browsing this forum: DuckDuckGo [Bot], Google [Bot] and 9 guests
