Accuracy

Posted: Fri Feb 05, 2010 5:50 am
by rgamage
I expect a peak at 515 and observed a peak at 516m/z. Is this close enough? What possible reasons can cause this difference?

Posted: Fri Feb 05, 2010 11:27 am
by Ron
Two possiblilities. 1. The sample has elements with large mass defect so monoisotopic mass is larger than nominal mass.
2. The mass spectrometer needs to be retuned to adjust the mass axis.
2 is most likely the case

I just thought of a third possibility. You didn't specify the type of instrument and ionization mode. If you are doing CI usually don't get molecular ion, you get molecular ion plus a proton.

Posted: Sat Feb 06, 2010 12:41 am
by Don_Hilton
You may be looking at something else. Give a bit more detail about what you are doign - instrument, compund, ionization thechnique.

Re: Accuracy

Posted: Sat Feb 06, 2010 5:08 pm
by thohry
If the instrument has passed a calibration (tuning), it can not be that big difference. You may try to re-tune the MS.