DETECTION LIMIT ON LCMS
I recently started working on the LCMS, and I have wondering what is most reliable way to determine the LOD and LOQ for my analyte. Using my calibration curve (standard deviation of the intercept method) gives really high values for the LOD and LOQ. Is there any other acceptable way of doing this apart from S/N? Thank you.
Hi Folarin,
sure there is a different way. Have a look: https://www.agilent.com/cs/library/tech ... 7651EN.pdf
Hi Folarin,
Hi Thomas,
How do I estimate the LOQ using this method? Thank you
Did you read the paper?
You make some injections (7-10) of a small standard and determine the standard deviation. You can then calculate a method detection limit by multiplying the SD with a t-Value (Students t for a certain 1-alpha and n-1 degrees of freedom).
Calculation example is given in the paper.
For a more detailled explanation of the method you could check the EPA document: https://www.epa.gov/sites/default/files ... 3-2016.pdf
The MDL is equivalent to a LOD. To get a LOQ, you can either just multiply by a factor (usually 3.33) or use another statistical criterion, e.g. the lowest spiked concentration that gives an acceptable RSD (e.g. <20%).
Hi Thomas,
I read it 😅.
It didn't really mention anything related to LOQ, and I just wanted to be sure. Thank you for your response.
In total there are 19 users online :: 2 registered, 0 hidden and 17 guests (based on users active over the past 5 minutes)
Users browsing this forum: Google [Bot], Semrush [Bot] and 17 guests
