Error column oven too slow response Scion GC
Hi all,
Our Scion 436-GC is having problems stabilizing column oven temperature, and the error 109 "column oven too slow response" is showed.
Anyone can help me with these problem?
Thanks a lot!
Sol
Sol
I don't know your instrument but when I've had this sort of thing happen on my Agilent systems, it usually means the flapper motor has failed. There's likely a flap of some sort in the back that keeps the hot air in or lets it out. It's not a difficult repair on a 5890/6890.
It could be an electronic, electrical or mechanical problem. Unless you have lots of time to spend troubleshooting it is probably worth asking a service engineer to call.
Thanks for your answers. I think it is the column oven temperature probe that is damage. I'll change it and tell you if we solve the problem.
Sol
Hi, I would like to know if changing the column oven temperature probe solved the problem.
Thanks!
Hi, sorry for the delay, I've not return to the site until today. Yes, it solved the problem.
Sol
Users browsing this forum: Google [Bot], Semrush [Bot] and 6 guests
