anyone can recommend any review on LH-20
Posted: Thu Feb 18, 2010 4:22 am
it is odd that people do not recommend use of mixture solvent,only use one pure solvent. how can they gurantee complete elute of analytes in sample?
Chromatography Forum is a public discussion group where you can post questions, news, or messages of interest to chromatographers everywhere, but you must be registered to participate (registration is FREE). If you are a registered user, please log in.
http://chromforum.org/
You can *never* guarantee complete elution in liquid chromatography.
Usually, but not always. That's why you will see "regeneration" procedures in the literature that recommend flushing reversed-phase columns with solvents such as methylene chloride.small molecules sample after SPE in C18 can be completely eluted out with up to 100% gradient organic solvent.
That said, Sephadex LH-20 is a different animal. It was never designed for use under HPLC conditions, and lacks the rigidity necessary to tolerate large changes in mobile phase composition (not to mention pressure!). That means that many of the procedures we take for granted with HPLC columns (gradients, high flow rates, backflushing the column, large changes in solvent types (so long as they are miscible!) will not apply.
If you have strongly bound material in your typical mobile phase, the best approach may be to unpack the column, clean the packing "batchwise" with an appropriate solvent, and then repack the column. What constitutes and "appropriate" solvent, of course, depends on what kinds of strongly-bound contaminants you expect.
I'm not sure exactly what you're looking for. This is a product that has been around for 40 years. A Google search for "Sephadex LH-20" turned up over 200,000 hits, including this one:did you see any good review paper on LH-20, i did not see any.
http://www.skkupharm.ac.kr/krlee/erp/er ... exLH20.pdf
There has to be *something* useful in there!
Why would there *not* be partition (or adsorption)? Minimizing "secondary" interactions with the stationary phase is probably the hardest part of doing GFC of proteins.i just do not know why there is partition mode in LH-20 chromtogram.
jiang295, you can look into attempts to explain HILIC to find out why there might be partition with "solid" stationary phases.