Resolution calculation
Dear all,
I know two equations for calculation of resolution:
- one considering retention time differences and peak widths
- one considering N, k and alpha
Does anybody know a site or article where the algebraic transfer from one into the other equation is performed/explained (step by step) considering additional equations such as k = (tR-t0)/t0 or similar?
Thank you so much
Florian
It's a nice question.
I have substituted the equation R = 1/4 *SQRT(N)*(k/k+1)*(alpha-1)alpha with the following:
N = 16(tR/Wb)^2
k = (tR-to)/to and
alpha = tR2/tR1
and it results in the other equation !
The first one is a calculated equation. This provides you the exact resolution you obtain from your chromatogram.
The second one is an approximate (theoritical) equation which shows how resolution varies as a function of other parameters you mentioned.
Thus I don't see why you would mathematically switch from one to the other.
In total there are 19 users online :: 0 registered, 0 hidden and 19 guests (based on users active over the past 5 minutes)