5973N MSD LCD screen
9 posts Page 1 of 1
I just picked my 6890N/5973N GCMS unit. When I plugged it in I noticed the LCD on the 5973 is not working. When I called the vendor he said "it didn't really matter, the screen go out and it only used for diagnostics. You will primarily control via the computer" - is this correct? Do I really need this to work if what he says is true?
GCMS newbie
Its true. I have a 5973 and the screen failed many years ago. Its usually not the screen, its the driver electronics in the very expensive smart card. Poor design I understand. Chemstation on your PC and the setup utility takes care of anything (e.g. setting 5973 IP addresses; updating smartcard firmware) you would be able to do with the the 5973 screen and controls.
We have four 5973 and two 5975 and I think two have working screens between them. I disconnected one when it actually began smoking lol.
Yes, they work just fine without them. Only thing I ever used them for was to monitor vacuum on the 5975 with the ion gauge, or turbo speed on startup.
The 7000MSMS and 5977 have more controls built into them, but you can still run them without even looking at the screen, so those really don't need them either.
I'll be the third. I have 2 5973N's and neither have the LCD working. The Chemstation talks to them just fine. They generate excellent data for me. It's been like this for years. Honestly, I don't remember the last time they actually worked!
I make a correction, the 5977 no longer has the screen. I just looked at our new one, I had not even noticed there was no screen now, that is how little I pay attention to the screens. :roll:
Thanks everyone for chimming in. Glad to hear that the 5973 MSD will work just fine with a non-functional LCD.
I have a 5973 with a non functioning Lcd screen and do not know the assigned IP address so Chemstation cannot see it to load the instrument top. Also have used the Agilent communications software to try and see if it would detect the MSD and it doesn't even see my 6890GC. Basically I need some ideas on how to find out what the IP address of the MSD without the use of the LCD screen and the Agilent software. Any help is appreciated
Open the command prompt in windows (windows R then type cmd)
type arp -a and hit enter
If you're lucky it will give you the IP address and MAC address of each device on your network.
The MAC address of your MS might be written on a sticker on the card in the instrument. But if not, you should be able to figure out which one it is by unplugging the MS from the network, using the arp -a command a second time, and seeing which IP address disappeared. You could also Google to find out what Agilent's MAC addresses look like (they might all start with the same few letters or something like that).
The instruments require a static IP for the Host, GC, and MSD. You may be able to alter them but I believe these are the default IP addresses.
5973 MSD IP 10.1.1.102
6890 GC IP 10.1.1.101
Host PC IP 10.1.1.100
9 posts Page 1 of 1
Users browsing this forum: Semrush [Bot] and 1 guest
