HPLC seperation problem
Hi All,
I had a problem with seperating acetylene and oxaloacetic acid with HPLC method.
Some one could help????
Someone may be able to help if you supply more information on how you are attempting the analysis and details on the problem you are encountering.
Hi All,
to clarify this problem,
the substrates I am going to seperate have identical structure and acidity (i guess).
I have tried ion exchange column, ion exclusion column, can't work this out.
Therefore , I would really need some help for separating them. Like which kind of column I would worth to try.
Thank you very much in advance.
What is your detection technique? My main concern is monitoring acetylene.
In terms of retention and separation your solutions is mixed-mode chromatography. Acetylene will retain by weak reversed-phase interaction and oxaloacetic acid will retain based on anion-exchange mechanism.
Here are few applications for neutral and acidic compounds:
http://www.sielc.com/application_165.html (vitamin C is acidic and Vitamin B2 is hydrophobic)
http://www.sielc.com/application_163.html
http://www.sielc.com/application_136.html
Contact me if you have questions.
You can also consider doing GC.
Vlad Orlovsky
HELIX Chromatography
www.helixchrom.com
Detection was done by a UV-Control detector at 210nm.
It is found that the peak of enol-OAA overlays with ADC
btw, does OAA's enol and keto form should compose two peaks in the HPLC measurement?