pressure abnormalities, air bubbles, inefficient venting
My problem has begun 30 days ago. When I turn on my HPLC I observe to low pressure and large pressure fluctuations. My pressure should be 9 MPa, but when I turn on HPLC, pressure is only 4 or 5 MPa and it fluctuate about 20-30%. Of course I start to purge the pump, but it takes me 1 or sometimes 2 hours. When venting is finally finished pressure is very stable, fluctuate less than 2% and pressure is 9 MPa. Earlier I didn’t have such problems, purging was necessary only after, for example, 10 day long brake in apparatus usage
I have on line solvent degasser.
Please help me resolve my problem.
The degasser needs mending.
Better yet: Toss it away and degass your mobile phase prior to use.
Best Reards
Thank you for your answer. But I have doubt.
If degasser is damaged, why after long pump purging pressure back to normal?
Well, then we’ll have to remove your doubts :wink:
In the beginning the air bubbles released in the pump heads and elsewhere in the system are very tiny. Over time they “find friendsâ€
In total there are 25 users online :: 0 registered, 0 hidden and 25 guests (based on users active over the past 5 minutes)
Users browsing this forum: No registered users and 25 guests
