Does anyone have a reliable method for cleaning up a fipronil product? We tried one using florisil and acetonitrile. However, we did not get a possible fipronil peak using the GC. Our MS is down, so...Really looking forward to a reply.
We haven't done by GC; are you using FID or ECD for that?
We assay for fipronil using HPLC, a test method we validated.
Our HPLC is not set up. Anyway we used a ECD method - we have not been able to validate it and our MS is down...But we were forced to get some results. Anyway the standards did show a consistant peak around 18mins or so.
QuEChERS dispersive SPE clean-up with magnesium sulfate and PSA works well when the Fipronil is in acetonitrile.
Don Shelly
Don Shelly Consulting, LLC
don.shelly@donshellyconsulting.com
Thanks Don Shelly
Could you please provide a specific method?
If you could please send me an email I will reply with a method. My email address is dshelly@unitedchem.com
Don Shelly
my personal email is allsaintsbaby@hotmail.com Thanks for the reply.