Hi all
I would like to ask if there is a list of commonly seen peaks (mobile phase components, atmos[heric components, clusters...) for LCMS, both positive and negative modes? Ionspray and APCI?
Any good links?
Thanks
Alp
try this article
Analytica Chimica Acta
Volume 627, Issue 1, 3 October 2008, Pages 71-81
Interferences and contaminants encountered in modern mass spectrometry
Bernd O. Kellera, , , Jie Suib, Alex B. Youngc and Randy M. Whittald
You may try this link and download the file called "background Ion Master List".
http://www.waters.com/waters/support.ht ... &type=SSPR
Hubert
Rad, Hubert,
Thanks for the links.
I found the Waters PDF useful and a search for
Interferences and contaminants encountered in modern mass spectrometry Bernd O. Kellera
led me to another useful file.
Exactly what I have been looking for.
Thank you!