epc oscillation
Posted: Mon Jul 26, 2010 1:08 pm
by chromtech
I have a 5890 series 2 that has been upgraded with an epc system. With the purge valve off it controls column head pressure fine. When the valve is opened for split injection the pressure oscillates above and below the set point. Is this a symtom of an eprom mismatch?
col. head pres: 12.6
split flow: 120ml/min
Posted: Mon Jul 26, 2010 3:19 pm
by Peter Apps
Check that the pressure feeding the GC is high enough, and that there are no restrictions in the lines.
epc oscillation
Posted: Mon Jul 26, 2010 6:14 pm
by chromtech
carrier pres. is @85 psi to the gc regulator, no kinks or obstructions.
Posted: Mon Jul 26, 2010 9:22 pm
by bhuvfe
what is the spread of the oscillation?
Maybe check for leaks at the inlet (starting from the septum).
Posted: Tue Jul 27, 2010 6:37 am
by Peter Apps
If you have a trap between the inlet and the back pressure regulator (I am assuming that you have the usual kind of plumbing for a split splitless inlet) it may be blocked or significantly restricted.
5890 epc oscillation
Posted: Wed Jul 28, 2010 12:46 pm
by chromtech
I have been all over the gc with a sniffer, can't find any leaks. Also the oscillation doesn't seem to start until the instrument is connected via HPIB to chemstation software. There is no trap, don't ask me why.
Thanks for any help in this forum, it is greatly appreciated
Posted: Wed Jul 28, 2010 2:58 pm
by Peter Apps
Check that all the settings in Chemstation match reality; column dimensions, carrier gas, pressure units, septum purge flow and anything else that I have forgotten. Also it can't do any harm to measure the actual flows with an independent flow meter.
epc oscillation
Posted: Wed Jul 28, 2010 5:10 pm
by chromtech
after substition found that the septum purge regulator was faulty.
thanks again
Posted: Thu Jul 29, 2010 6:13 am
by Peter Apps
Thanks for the feedback