Ammonia by HPLC?
Is HPLC a good way to measure ammonia? What kind of detector would work? RI?
What about GC? Would that be better?
Or is there a better method altogether.
I'm monitoring ammonium usage during a biological reaction.
Thank you for pointing me in the right direction!
H_H
Hello H_H; ammonia can be analyzed either by HPLC (using ion exchange) or GC; but sometimes and depending of your concentration; steam destillation and Titration could be easier solution; if your problem is sencibility you can also use Selective Ammonia electrodes or colorimetric method; it is necesary to have in mind that these tree methods are sencitive to volatil amines also but work easier and faster.
Hallo HPLC_Help
oscarbal is absolutely right. Ion Chromatography, titration or ISE all work in their respective concentration ranges and sample matrix conditions. For hgher concentrations I would go for Titration / ISE. For lower concentration and/or amines around, IC will be the preffered method.
Dr. Markus Laeubli
Manager Marketing Support IC
(retired)
Metrohm AG
9101 Herisau
Switzerland
Also can use the Nessler reagent and do it spectrophotometrically.
Try Inertcap column for volatile amine u can get good responce not only for ammonia but other volatile amine also on GC-TCD
Praveen
here is method for analysis of ammonia using Primesep mixed-mode column and ELSD detector. You can analyze non-UV active amines using this approach:
http://www.sielc.com/compound_005.html
Contact me if you have questions