pH effect on Evaporative Light Scattering?
Hello:
Recently I've been told by a colleague that mobile phases with a pH above 8 would be a problem when using ELSD.
Does anyone know if this is true, why that would be the case (theory) and if they've seen any literature discussing this?
thanks
For ELSD (like for LCMS) you need to use volatile buffers, such as an amine in your case.
So, Ammomium Hydroxide wouldn't be good with ELSD?
It should be fine to use ammonium hydroxide, it is a volatile base. Sodium hydorixde wouldn't work but the ammonia should be no problem. Just out of curiosity what type of separation are you using w/ the ELSD? I have been working on an AEX method to transfer from UV to ELSD and I am interested to see what others have come up w/ for this separation w/ the ELSD.
Ryans17,
There is no correlation between the mobile phase pH and what is or not compatible with the ELSD. This detector just need a volatile mobile phase which is independant of the mobile phase pH...
Maybe the folklore about not using buffers above pH 8 came from a case where high pH was dissolving a column packing and the silica was showing up as a signal. Or the high pH was dissolving something in the detector. As others have already stated, there is nothing inherent about pH 8 that is bad.
Bill Tindall