RSD of <2%. Is that possible with LCMSMS?
Okay, I am going in circles so I thought I might as well ask.
From my other postings, it is obvious I had been working on LCMSMS analysis on melamine in milk samples.
I am told I must attain less than 2% RSD. I had tweaked all that I can and is still getting 3-4% RSD at a minimum. Sometimes, it gets as high as 12% but mostly less than 5%.
Is a <2% RSD even possible with LCMSMS, when analysing actual samples???
When I was doing Bioanalytical assays by LC-MS-MS we had our SST criteria set at RSDs <=5% as our default.
Good judgment comes from bad experience, and a lot of that comes from bad judgment.
While I understand the grave consequences of finding toxic levels of melamine in milk samples... may I ask why, from a practical standpoint, less than 2% RSD would be necessary? If I knew with 95% certainty that a milk sample contained a certain concentration of melamine plus or minus 10%, would that 10% error have any bearing on how toxic the milk is?
When I was working in a routine lab, the default was set at <5-10%, depending on the type of samples.
Now I am working in a metrology lab, so they expect high precision and accuracy.
With GCMS, I am able to attain RSD <2%. However, no matter how hard I try to tweak the parameters, the minimum I obtained for LCMSMS is 3-4%.
In total there are 22 users online :: 0 registered, 0 hidden and 22 guests (based on users active over the past 5 minutes)
Users browsing this forum: No registered users and 22 guests