Difference between LOQ and LLOQ
Could someone explain what the difference is in LOQ and LLOQ? Is there a difference in determination? We use LOQ at S/N >10.
THanks,
Wanda
LLOQ is a term usially associated with Bioanalytical methods, in their guidance document http://www.labcompliance.de/documents/F ... 252fnl.pdf the FDA define it as:
Lower limit of quantification (LLOQ): The lowest amount of an analyte in a sample that can be quantitatively determined with suitable precision and accuracy.
it is not a value determined in the same way as LOD but more a determination of the practical working limits of the method in relation to acceptance criteria.
For bioanalytical methods the lowest calibration/QC STD is referred to as the LLOQ STD.
Good judgment comes from bad experience, and a lot of that comes from bad judgment.