Hi everyone!
Im trying to analyze lysine in an RP column(C18) in a PDA detector but still no peak in the elution. can you give me any suggestion? thnx everyone!
I think you'll need to find another detection technique (not UV absorption based) or derivatize the lysine.
You could probably see something below 210 depending on if your mobile phase is transparent down there. i.e. really clean ACN and water with no additives. Sensitivity will still be extremely poor.
Hi dude,
Lysine is an amino acid. Generally amino acids are not detected with UV. It requires derivatization. If you search on Google for amino acid derivitisation methods you can get some and may helps you.
Best regards
Babu
you will not have too much retention on C18 column without IP reagent.
you can see lysine in UV at 200-215 nm. here is few applications including one with UV detection:
http://www.sielc.com/compound_057.html
contact me if you have questions
Dude! Where's my Lysine?
Sorry, couldn't help myself!
:lol:
Have you tried using pre-column derivatisation with O-Phthalaldehyde (OPA) in a borate buffer, giving UV absorbing iso-indoles. Run this on a Hypersil Gold column at 40 C, using a binary gradient . Use a binary gradient of 40mM NaH2PO4/Na2HPO4 (1:1) buffer pH=7.8 and AcN/MeOH/H2O (45/45/10 v/v/v). Detect at 338 nm. Lysine peak should elute at around 15 minutes.
Kind regards,
Ade Kujore
Marketing
Cecil Instruments
Cambridge
United Kingdom
email:- ade.kujore@cecilinstruments.com
telephone:- +44 (0) 1223 420821
web site:- www.cecilinstruments.com
Registered Number 909536
8 posts Page 1 of 1
In total there are 18 users online :: 2 registered, 0 hidden and 16 guests (based on users active over the past 5 minutes)
Users browsing this forum: Bing [Bot], Semrush [Bot] and 16 guests