Fermentation Broth components by Headspace GC Development
Posted: Sun Nov 14, 2010 4:06 pm
Hi All, I'm new here and have just started a new job.
I need to develop a Headspace GC method (using the Agilent HS) to analyze fermentation broth for solvents produced during process development using either an FID (preferably) or an MS detector.
I've done a lot of headspace work for residual solvents for small molecule drug delivery development in the past. These had very well defined relatively simple matrices and external standards always worked well.
But I have not had to deal with a messy matrix like broth before. Also the levels I'm looking for here are in a much broader range ( 250PPM to 25%) to follow the fermentation process.
For HPLC work, I centrifuge the broth and analyze the filtered supernate, which gives analyte concentrations in the supernate and not in the broth itself.
For HS-GC, in theory I should be able to just add 'raw' broth to the headspace vial and test that and get concentrations relative to the volume of added broth. Doing that, I realize it is highly unlikely I could get decent results using external standards.
To get the best results I would have to do some other method of quantitation.
The thing is they also want results as close to real time as possible. I think I can develop an isothermal method for the analytes, so how i do quantitation can affect turn around time significantly.
With the column I'm going to use my isothermal GC run time may be under 5 min and should almost certainly be under 10 min. I can always run standards early in the day before samples start coming in, so it's the per sample analysis time that matters most.
The options for quantitation that I can see are:
1) Standard addition:
Potentially most accurate, but slow (multiple injections per sample) and i don't know how well it would work for a broad concentration range.
2) Multiple-HeadSpace- Extraction:
Gets rid of matrix effects but slow (multiple injections per sample). Also I'm not sure how accurate it is in practice (I have never used it)
3) Full Vaporization:
Potentially fastest because vial equilibration time should not slow things down and could be a single injection per sample. It should also either eliminate or minimize matrix effects.
But I don't think I could accurately sample small volumes of raw broth because of the solids present.
4) Internal Standard
Potentially second fasted after Full vaporization as it would use a single injection per sample.
Theoretically not as accurate, but I MIGHT be able to get away with an internal standard (or two ISes) if the internal standards are similar enough chemically to the analytes to have partition coefficients affected to about the same degree as the analytes by the matrix.
Theoretically quantitation would not be as accurate as the above methods, but might be good enough practically.
If I work with filtered supernate then Full vaporization becomes an option I think,but I still worry about the precision and accuracy reproducibly of adding such small volumes for lots of samples.
With the supernate at least, maybe there is even one more option:
5) Internal and maybe even simple External standards with dilution.
Essentially try to dilute out the matrix effect. If I have enough sensitivity I could try taking say 50 or 100 ul of filtered supernate and adding 0.5 or 1 mL of Water or dilute internal standard in water to a 10 mL HS vial. That might give good enough results.
I'm looking for the best way to get started on this method development, which also needs to be done ASAP.
So am I wrong in any aspect of my analysis of the situation? Am I missing something? Any advice/thoughts on what to try first?
Thanks for any help.
- Karen
- Karen