Silica vs Wax Column
I am trying to detect propylene glycol in fuels. As of right now the PG is eluting a min into the fuels retention zone (its a broad section). The boiling points are similar but as the fuels are not polar at all and the PG very polar would a change in column help? We are currently using a 30m fused silica column.
When we assay for propylene glycol in consumer products (not fuels) we use trimethylsilyl derivatization and a non-polar column.
We've also done some glycols directly on 624 columns, such as Rtx-624.
The important thing about the column is not that it is fused silica, but what is coated on the inside; the stationary phase. Since you seem to have a boiling point separation (or lack of) at the moment you are probably using a non-polar phase. A polar phase will retain the glycol more in relation to the non-polar constituents, this may or may not give you the separation that you need. Have a look on the column manufacturers' web sites, and ask them for advice.
Keep in mind that underivatized glycols are susceptible to adsorbtion, so peaks are often bady tailed.
In total there are 31 users online :: 2 registered, 0 hidden and 29 guests (based on users active over the past 5 minutes)
Users browsing this forum: Ahrefs [Bot], Semrush [Bot] and 29 guests
