Negative peaks at tail of analytes suppressed IC system
Posted: Thu Dec 09, 2010 1:57 pm
We are seeing negative peaks occurring on our Metrohm suppressed Anion IC system. We have tried a new column, have cleaned and checked our injection system, tried multiple sources of eluent and standards, and cleaned the suppressor system according to metrohm spec.
The negative dip occurs at the tail of most of the analytes we are separating (not SO42-, perhaps because it is a broad peak?). We only integrate for Cl- and SO42- and it is interfering with our Cl- calibration curve. The problem persists in all of our standards/ working samples and gets more pronounced with higher concentrations. Our working samples consist of Denatured Fuel Ethanol (no dilution) and Antifreeze (100:1 dilution). At this point we have tried everything that we can think of and are at a complete loss.
Any suggestions? Thanks!