Baseline Drift & Noise
I have searched many references for procedures and specifications for detector noise and drift.
The general consensus seems to be equilibrating the system in 100% degassed water (deionized, WFI, etc.) and recording the trace over 5 minutes - 1 hour. Noise = difference between the maxima and minima every 30 seconds. Drift = absorbance maxima and minima difference over the 5 minutes to get AU/minute or AU/hour.
Are there any better procedures around?
Also, I can not seem to find any industry standard specifications for these tests. Since we have several different types of HPLC systems, manufacturers, and detectors, I do not think it wise (or necessary) to go by the manufacturer's tight specifications. I would like to assign a generic, reasonable specification for all the systems to calibrate against.
First, you may want to check ASTM. (I would, but www.astm.org is inaccessible right now.)
Most chromatography data systems have a built-in algorithm for calculating noise and drift. If your facility has standardized on one software, just use (and document) that as your algorithm. If you have a few odd instruments, you can export/import the data to your usual software for the calculation.
For instance, the Dionex software computes the regression line for a section of baseline. The slope is the drift and the noise is the sum of the maximum and minimum excursions from the line. Other software uses the RMS.
Mark Tracy
Senior Chemist
Dionex Corp.
