Hey guys,
I was wondering, is it possible to separate methanol and isopropanol in the gas phase on an agilent db5-ht?
I don't think it is, since it is not a polar column? I see a good methanol peak which flies through after about a minute, but with the isopropyl it just doubles the peak pretty much, i don't get separation.
I am doing headspace analysis with gas tight syringes. I have made standards at 0.01%, 0.2%, 0.5% and 1% methanol and get a nice linear curve and r squared of 0.994, but my response factors seem out of whack, so i am trying to use an internal standard. However I am limited to only using a db5-ht for the time being. Should I try for another compound to use as the internal standard?
I am taking 2mL sample, heating in a 12 mL vial to 80 deg C for 5 minutes, and injecting 200uL of headspace directly on the column. Helium at 3mL/min, injector port at 70degC , oven isothermal 70 deg C for 6 minutes. Should I try a cooler oven temp and then ramp up temp slower? I am also trying to do these tests in the shortest time as possible.
Thanks!
Film thickness is everything and of course, the length and ID of your column.
If you are using a 0.53mm ID 15 meter column with 0.16µm film, BAD CHOICE.
You may wish to increase the temperature of your injector to 120°C - 140°C
Lower your oven to 50°C isothermal.
Use a low volume injection liner and inject slowly.
I think if you check the Supelco or Restek or Alltech web sites you will find chromatograms of methanol and ipa on a methyl silicone phase column for comparison.
best wishes,
ROd
Presuming that you made you standards in a matrix that matches you samples, and since you get a good (enough ?) linear fit, you might not need an internal standard at all. I'm not sure what you mean by "response factor" - methanol gives a lower FID signal than the same mass of hydrocarbon, and lower than the same mass of isopropanol, because it has proportionally less C-C and C-H in it.
Given that you cannot change the column, low oven temperatures are your friend.
thanks for the help guys!
Ok on the math side I was under the assumption that to determine the quantity of an unknown analyte, you have to first run a standard and get the response factor= (standard peak area)/(concentration of analyte in standard). This is a single point standard. I am using a concentration near to what I think it should be of the unknown analayte.
Once i get the response factor, I take the area of the unknown sample's peak and divide it by the response factor... But when I look at the response factors from 0.01% to 0.5% they all give different percentages for the unknown analyte amount. Shouldn't they all be roughly the same? I am seeing differences 0.2%, 0.4% and 4% all for the same sample.
So I thought I would need an internal standard to correct for the variation in headspace sampling as well as the GC. I want to test how reproducible my headspace sampling methods are as well as get an accurate number of %methanol.
On the sampling side I am using our biodiesel as my matrix its water washed multiple times and heated to 100 deg C for 30 min. I then spike this with methanol. I do all injections with a heated needle and prime it 3 times before piercing the headspace vial, and then prime it 3 times in the vial itself. I then inject.
I have tried your method chromatographer but I am getting random peaks and noisy baseline when i heat my inlet to 120 degC, could be contamination/septa bleed?
Just so that I am absolutley sure - your are trying to measure how much methanol there is in you biodiesel ?
Given that you have a linear fit of peak area vs amount of methanol the response factor must be the same at the different levels of your calibration. If the response factor (signal/analyte mass) changed with analyte mass the fit could not be linear.
Since you already have a 4-point calibration, calculate the equation for the line of best fit (peak area = x times analyte mass + y, where x is the slope of the line and y the intercept), then run a sample and calculate analyte mass from that equation.
AZbiodiesel,
The HT series have almost no film thickness to them at all. This is so they can go to very high temperatures to measure high MW hydrocarbons. Plus, they're non-polar phases. I would say that the only way you are going to separate meoh and IPA is with cryo at something like -20. Thus, your other idea of picking some other constituent other than IPA might work but, to get your separation, you are going to end up with a component that is not anywhere near the volatility of meoh. Is that a good choice for IS, I don't know.
For what it is worth, you could consider an external iso-thermal oven (a valve oven works great for this) for meoh analysis on the same GC.
Best regards,
AICMM