Page 1 of 1
GC heated zones are not working
Posted: Wed Jan 26, 2011 1:20 pm
by JI2002
We have a 5890 series II GC with TCD and FID, when the GC was started today, all the heated zones including injection port, oven and both detectors stay at room temperature. There is no fault messages, it only shows oven temperature not ready, injector A temperature not ready, detector A temperature not ready and detector B temperature not ready. Powering on and off the GC didn't help. Rebooting computer didn't help. Fuses on the board checked out OK. Any ideas?
Re: GC heated zones are not working
Posted: Fri Jan 28, 2011 2:31 pm
by JI2002
Agilent service representative came in and fixed the problem by replacing the motherboard.
Posted: Fri Jan 28, 2011 3:59 pm
by Peter Apps
Thanks for the update. I'm glad that it's fixed.
Posted: Fri Jan 28, 2011 4:07 pm
by Consumer Products Guy
So the 5890 motherboard was still available....
Posted: Fri Jan 28, 2011 6:11 pm
by JI2002
The service guy had only one motherboard for the system. He has been in the business for a long time, so he has parts for old GCs.