Waters 2695 battery change
2 posts Page 1 of 1
We have injection errors on our 2695. The service engineer suspects the battery. We have a spare battery.
Is there any guide on how to change the battery?
Having a service engineer drive a few 100 km just to change a battery seems a little bit decandent to me
Alex
Problem solved. Just for the protocol:
remove two screws on the left side,
slide the cover backwards
find the battery in the upper left quater
change it
close everthing
run a seal adjust