NOx by GC?
Is it possible to analyse the concentration of NO in a sample by GC using FID detection? Or do I need to use TCD?
Barbara
Your best choice would be a Helium Ionization detector by AIC, or Valco, or Gow-Mac.
What is several thousands of dollars when you need the best, most accurate, and LOWEST numbers possible?
Want more?
A Mass Spec for tens of thousands of dollars?
Low temperature porous polymer packed column is the best column choice, but get the best packed, not the cheapest. It does make a difference. Ask for cleaned packing to be used when you order. Pay the few dollars more for much improved performance and fewer and smaller voids. Yes, into our lives a few voids must fall.
Have a good week,
Barbara,
Depends a lot on the level. If hundreds of ppm then a TCD would work fine I think. If low level or really low level then I would recommend a PID or an AID (argon ionization detector.) (The standard disclaimer applies here, I make both of these detectors.) With an IP of about 9.2 both of these detectors would have good sensitivity for low level NO. I am not sure the impact of the window in the PID but this compound has been done by AID and I can post a chromatogram if you would like. You will need to watch out for CO2 as it may be a potential interference. The other thing to be aware of is reactivity and this is one of the big reasons to pay attention to column quality.
In this particular case, I would not recommend an HID since you will have to have very good separations since the HID will see everything else you throw at it. I often state that an HID is like a TCD on steroids, it will see everything but with much stronger signals (lower detection limits.)
Best regards,
AICMM
Thanks so much for both your replies! Much appreciated.
Barbara
Users browsing this forum: DuckDuckGo [Bot] and 1 guest