I have a paper by restek that lists about 280 pesticides that can be done by lcesimsms - some negative but mostly positive ion modes. Are there any other freebies out there?
vestel b. shirley, president
betves inc
166 norwood drive
reidsville, nc 27320
The CRL Data Pool (requires registration)
http://www.crl-pesticides-datapool.eu/
A downloadable set of acquisition parameters for ESI-MS/MS (API 2000):
http://www.bfr.bund.de/de/bfr_daten_fue ... -5831.html
thanks.
I am actually trying to determine if I can screen for the possible crop chemicals by syringe/lc/esi/ms/ms for a specific crop in our state (or any state that has a good database). It sounds do-able (detection limits notwithstanding). Given the transitions and instrument conditions database(s), I can write the program. You've been very helpful.
again, thanks.
betves inc
166 norwood drive
You can also try asking your instrument manufacturer. One of ours supplied us with a hefty file containing transitions and collision energies etc for a huge number of pesticides and other compounds. Of course you'll probably have to do a little fine tuning for your methods.