HP 6890/5972A GC-MS: "cannot bring up GC" error
Posted: Fri May 27, 2011 2:50 pm
by sirpatty79
Hi everybody,
I am using an HP 6890/5972A GC/MS w/ Chemstation GC1701BA version B.01.00. The software was recently reinstalled and now will not connect to the instrument. The interface is an Agilent 82350B PCI GPIB. Thanks in advance for any help you can provide!
-Pat
Re: HP 6890/5972A GC-MS: "cannot bring up GC" error
Posted: Fri May 27, 2011 3:08 pm
by Consumer Products Guy
Did you configure the GPIB board as "hp82341" even though it is an 82350B ? You need to. And change the bus address to 30. You need to.
Posted: Fri May 27, 2011 6:11 pm
by sirpatty79
Thanks very much, that fixed it. The problem was an incorrect SICL name and bus address. You guys are a great help!
Posted: Fri May 27, 2011 11:31 pm
REALLY????
Well, great - Because I have no idea what a SICL name is, and to me a bus address is where your old VW Type 2 is parked. And isn't a router used in the woodshop?
Posted: Thu Jun 02, 2011 1:32 pm
by gcguy
Isn't software a type of comfortable shoe?