how to optimize to obtain optimal MS conditions for charanti
hi,
i'm running a charantin standard to optimize MS conditions. charantin's molecular weight is 576. the problem is i can't see any ion has this m/z value. some ions appear in MS spectrum is 613.381, 429.234, 174.897, 196.879 (high intensity); 379.126, 429.234, 522.592 (small intensity). the mass of 613.381 has the highest intensity in the spectrum. these ions appear in the spectrum at the value that i don't expect. because charantin has a glucose group in its structure. i think it will appear the mass of 577 and 413 (release the glucose group has m/z value at 162). i would like to know what i had done wrong. i'm using MicrOTOF-Q II MS instrument of Bruker.
Some value of MS source:
- Source:
end plate offset: -500V
Capillary: -4500V
Nebulizer: 0.5 Bar
dry gas: 5.0l/min
Dry Temp: 180 C
- TRansfer:
Funnel 1 RF: 200.0Vpp Funnel 2 RF : 200.0Vpp
ISCID Energy: 0 eV Hexapole RF 200.0Vpp
- Quadrupole:
Ion energy: 20 eV Low mass: 100m/z
- Collision Cell
collision energy: 20eV Collision RF: 150.0 Vpp
Transfer time: 60.0 us (microsecond) Pre puls storage: 4.0us
- Detector:
source: 0V
Please help me. i need your help!
The mass 613.381 may be the common ESI- adducts M+K-2H (M+36.9486). I am not sure without having the exact mass of charantin.
You may want to run a CE ramp to find the right CE you need.
thank for your reply!
charantin is a mixture of two compounds: sitosteryl glucoside (C35H58O6) and stigmasteryl glucoside (C35H60O6). I think the mass of 613 is sitosteryl glucoside (M+K). however, i can't see the mass of 615 (stigmasteryl glucoside).
hi thientham,
i'm doing a research about charantin. can u give me suggestion for any supplier companies which can provide me charantin hplc standard 99%?i find it difficult to find the charantin hplc standard.
thanks.
The value is too high in my thought, the compound might be broken under such high energy.
Another thing is that you should always have control, e.g., a blank run, to be certain of that the signals are from your sample instead of the system.
hi thientham,
Try http://www.phytolab.com