I have a problem with a HPLC Waters Alliance 2690,
I miss It's Software and Interface(IEEE 488),and I use of Windows XP and I Can Prepare Mellennium or Empower Software,
Would You tell me I must prepare Which model interface and software that work in Windows XP,With a suitable Price?
P.Chegini
The IEEE interface requires a BUS-LAC/E board to be installed in a computer running Milennium or Empower software (teh board and computer are generally packaged together as a LAC/E, but you must be sure you find a BUS LAC/E as not all LAC/Es come with the BUS-LAC/E board installed). Consult someone at Waters.com for pricing.
Thanks,
DR
Hello Evrybody,
To run your Waters Alliance 2690 with a good software you will need:
1. A desktop computer with PCI slot and Windows XP professional as operating system.
2. A PCI bus LAC/E card.
3. Empower 1, 2 or 3 software.
4. Upgrade your 2690 firmware to 2695 firmware version 2.03.
If you want to get a good software with full control over your system we can help you. We have all that you need. Send us an e-mail to info@greenlab.ws