HPLC with electrochemical detector
2 posts Page 1 of 1
hello
I had a topic about electrochemical detection but anyanswers
now can you tell me about the mode used in EC detection: DC, PAD and SCAN modes
in whitch case we use each one
the difference between them
how to construct a voltammogram: steps and method
thank you
Dear sofiane
DC = direct current is the basic version where a stable potention is applied to the electrodes. The measured current is proportional to the concentration of the analyte. Some reaction products my poison the electrode surface. In this case
PAD = Pulsed Amperometric Detection is applied. Here an oxidative and a reductive potential is applied after the analytical potention to renew the electrode surface.
Scan Mode: A potential scan is done to get the optimum working potential.
See also 871 Bioscan Manual.
Dr. Markus Laeubli
Manager Marketing Support IC
(retired)
Metrohm AG
9101 Herisau
Switzerland