What is the cause of this baseline step
22 posts Page 1 of 2
What would cause a step change in the baseline like this?
An integrator timed event to the baseline or detector.
best wishes,
I had similar problems at the University using UV laps well past theír actual life time , went away when we finally invested in a new lamp.
If it is reproducible then I would agree with previous poster.
It is absolutley electronic in origin.
Petrus Hemstrom
MerckSequant
Umea, Sweden
It looks like wavelength change to me - in accordance to what Rod suggested above.
Please review your detector method/parameters.
can be equilibration/saturation of ion-exchange/mixed-mode column with not fully transparent buffer/additive.
Vlad Orlovsky
HELIX Chromatography
www.helixchrom.com
if you make a step gradient then you can see such behavior in the baseline
if you change the wavelength in mid run and do not do an auto-zero as well then you can get this behavior as well
If this happen frequently at different time points, I am also think the lamp is bad.
The baseline between 3 and 7 minutes is also interesting, is this corresponding to a pressure fluctuation or is this a mixing problem?
Hi Klaus,
How does pressure fluctuations influence the UV signal?
Obviously the change occurs at a specific time, so I would look at the timed program. What is it that you have changing at the distinct time? Looks most likely to be a change in attenuation or something of that nature.
Hi Klaus,
Pressure instability means flow instability which does affect the UV signal
A. Carl Sanchez
Pressure fluctuation e.g. rise does not equal flow rate increase for instance. It could happen - and it happens all the time - that the gradient mixture changes towards more viscous nature. An example could be from watery to IPA containing eluent.
The noise early in the chromatogram is typical of a pump problem commonly seen with binary pumps when one pump (A or B) is not functioning properly
A. Carl Sanchez
You mean mixing problems. Could be. But nothing to do with the pressure itself.
Pressure and flow are correlated. Pressure instability will give flow instability.
A. Carl Sanchez
Not so! But the other way around!