HELLO guys
acording standard ASTM:D4747 i want to determine unreacted monomer content of latex(acrylic latex) by GC. BUT i've some problems,in procedure of this standard wrote: if the composition of the latex is not known, a preliminary analysis must be performed by diluting approximately 0.5g of latex with approximately 5g of warter ond injecting a 2microlitr into GC column.then by retention times to identify the monomers present. MY QUESTION IS:IS THIS SUSPENSION(0.5G OF LATEX + 0.5G OF WATER) DANGER FOR MY GC COLUMN? OR I HAVE TO FILTER THIS SUSPENSION? MAYBE IF I FILTER IT I'VE MANY ERRORS TO identify MONOMERS, my model of GC is:agilent 6890 N .
please help me.. thank you.
best regards.
HELLO guys
... ond injecting a 2microlitr into GC column...
Hope it is not on-column injection but into the liner.
What type of injector is quoted in ASTM:D4747 ?
Water injection is not a problem, assume it is with a good split. If you do not want to filter – centrifuge. Head space/SPME may be useful too.
"If your experiment needs statistics, you ought to have done a better experiment." Rutherford
You should check the volume of your liner and the volume 2 µl water takes at the injection temperature. I have an Agilent GC and the split/splitless liner has a volume of 900 µl. 2 µl water injected in an injection port at 250°C and a column head pressure of 100 kPa will give you a vapor volume of 1190 µl. In general you want no more than 2/3 of the liner volume as vapor. Under the above conditions the injection volume should be 0.5 µl maximum.
Gilbert Staepels
Ideas mentioned in this note represent my own and not necesseraly those of the company I work for.
In fact mona_mh is not afraid of water but rather of latex suspensed in water which was clearly articulated in capitals in her original post.