Varian 450-GC (TCD/FID)
4 posts Page 1 of 1
Dear Brothers
I have a Varian 450 - GC model Galaxie (TCD/FID) when injecting gas Hydrocarbons sample by manual , the device does not give true analysis , I used several parametere and changed the carrier gas ( H2 , He, N2) but no gives the true analysis
I use the packed column ( 30% DC-200/500 OP-AW 60/80 S.N 610060616 but I do not have the parameter of this column
I need your help
Are you trying to measure the BTU value of natural gas?
Yes I measure natural gas but no result
Do you not know HOW to perform the calculation? Your need is unclear.
best wishes,
4 posts Page 1 of 1
In total there are 9 users online :: 2 registered, 0 hidden and 7 guests (based on users active over the past 5 minutes)
Users browsing this forum: Google [Bot], Semrush [Bot] and 7 guests