spectra system P4000 pump solvent leakage

Posted: Wed Nov 09, 2011 2:30 pm
by parkerpal
whenever I start the hplc and begin running my method, the pump head begins to leak. maintenance was done on the pump recently, cleaning and replacing seals and pistons, and ever since then it leaks whenever I run my method. I removed the pump head and piston holder and made sure it was put together correctly and closed tight but that didn't make a difference. Any suggestions would be appreciated.

Re: spectra system P4000 pump solvent leakage

Posted: Wed Nov 09, 2011 3:52 pm
by tom jupille
You've already checked what would have been my first and second suggestions:
- are the seals installed correctly and
- is the piston scored.

You may have just gotten a bad set of seals (or inadvertently damaged them during installation).

Re: spectra system P4000 pump solvent leakage

Posted: Wed Nov 09, 2011 4:13 pm
by parkerpal
I have a few extra seals so maybe I can replace them and see if that helps. Thanks.