The default display with Agilent's CS E.02.xx on our 5975 is one digit past the decimal place. e.g. m/z 82.1, 181.2, etc.
To be consistent with the rest of our GC/MS's in the lab, I'd like this to round to the nearest whole number. I can't figure out how to do this. There is a supplied macro that creates a registry entries for both chromatogram and spectrum labeling, but it doesn't work.
Anyone know if this is possible to set the labels to whole numbers?
I had to ask someone.
In the Command Line, type MZDECIMAL 0
then select the Execute button.
Regards;