Tributylamine
does anyone have any advice/protocol for removing tributylamine (m/z 186) from a nanoLC/QToF system?
many thanks
Clive
Are you sure it's tributylamine? You would have to have had it in significant quantities.
Use an acidic washout solution - it'll help dissolve the TBA and carry it out of the system.
Mark Krause
Laboratory Director
Krause Analytical
Austin, TX USA
many thanks
Clive
Trace of TBA (186amu in +ESI) is very difficult to completely remove (after working with TBA).
1. Wash your HPLC with 0.1%v FA in water, then 0.1%v TFA in MeCN, then MeOH and water (approx. 50-100ml each). In some cases, you need to replace spare parts (degaser, capillaries, in-line/inlet filters etc.).
2. Wash injector/AS. In some cases, you need to replace spare parts (seals, syringe etc.).
3. Wash ion source.
P.S.: I don't use TBA anymore :)
Thanks for all the advice. We were using tributylamine as an ion-pairing reagent for the analysis of nucleotides, not knowing the sticky nature of the compound.
will try the advice and report back.
thanks
Clive
In total there are 21 users online :: 0 registered, 0 hidden and 21 guests (based on users active over the past 5 minutes)
Users browsing this forum: No registered users and 21 guests