CO2 & Air separation

Posted: Wed Dec 21, 2011 2:19 am
by neo_jm24
parameter setting on GC 2010 plus using VB-1 & PDHID for the separation of CO2 & Air???

Re: CO2 & Air separation

Posted: Wed Dec 21, 2011 3:02 am
by chromatographer1
describe the VB-1

Re: CO2 & Air separation

Posted: Wed Dec 21, 2011 6:57 am
by neo_jm24
VB-1 it is a ValcoBOND capillary column and is 100% dimethylpolysiloxane

Re: CO2 & Air separation

Posted: Wed Dec 21, 2011 7:46 am
by chromatographer1
You give no ID and film thickness.

Even so, it is very difficult unless subambient temperatures are used to separate CO2 and air using gas-liquid capillary columns.

A 8 meter 25% methyl silicone on Chromosorb P packed column can separate CO2 from ethane and propane even at 60C (140F) and air separates from methane well.

This is not possible using the bonded capillary column you mention.

best wishes,

Rod

Re: CO2 & Air separation

Posted: Wed Dec 21, 2011 8:01 am
by neo_jm24
Thank for you quick reply.

Currently I am using VB-1 capillary column (100% dimethypolysiloxane) and is 30m x 0.53mm ID x 5.00um Film. just wanted to separate the CO2 from Air.

Re: CO2 & Air separation

Posted: Wed Dec 21, 2011 8:05 am
by chromatographer1
Drop your temperature to -40C or so and they may separate. I do not know for certain the temperature required for that column. Of course, if you use a porous polymer PLOT capillary column it should be no problem even at 80C.

best wishes,

Rod

Re: CO2 & Air separation

Posted: Wed Dec 21, 2011 8:13 am
by neo_jm24
Thanks Rod, I am currently trying it now. I am new on GC specifically on system GC.