Square peaks on FID and tCd
I have 7890A GC with Front FID; Back FID and Aux TCD.
Initially, I used a gas composition of CH4        1
C2H4      1%
C2H6      1%
CO          1%
CO2        80%
N2         15%
O2          0%
H2        1%
Ar           0%
And this gave me nice peaks
I changed the composition to
CH4        10%
C2H4      10%
C2H6      10%
CO          10%
CO2        10%
N2         15%
O2          10%
H2        10%
Ar           15%
And now I started getting square peaks. Can someone help me.
Thanks.
oh come on.
No column or sample size is given.
Let's discuss after you can share these essential details.
I guess you might be thinking column overload.
Share?
fshah24,
Might I suggest you also post a chromatogram? FID's are extremely linear and that goes for the the electronics as well so I would very much like to see your chromatogram.
Best regards,
AICMM
Off scale peaks ?
Yes, Peter, but without the additional information we are all just guessing.
I hope fshah24 complies.
In your first injection the CO2 is 80% and in the second injection it is 10 %. In your second chromatogram the peaks are square, meaning amp is saturated, i.,e your CO2 also would have saturated, if this is the case then in your first chromatogram 80 % CO2 must be square i.,e saturated. To solve the problem, (1) attenuate your signal using software.
Vajjara
