Page 1 of 1
Any recomandation which liner to use for pesticide analytics
Posted: Wed Mar 14, 2012 12:01 pm
by Mr.Brown
Hello everybody.
We are starting with pesticide analytics by GC-MS (Shimadzu)
and i was wondering if anyone could recomand liners for this purpose.
I am thinking about SKY liners from Restek
Thanks for your help
Re: Any recomandation which liner to use for pesticide analy
Posted: Wed Mar 14, 2012 12:09 pm
by Hornet
Sky, there is no other answer.
If you tell us the instrumental conditions you want to use, we can think about the type.
Posted: Fri Mar 16, 2012 7:52 am
by Mr.Brown
We will use a OPTIC-4 Injection Unit for large volume injections (100 µL)
Therefore we will go for PTV injection.
We will use a Zebron MultiResidue-1 (phenomenex) column
Posted: Fri Mar 16, 2012 8:10 pm
by betz.jason
I use a ZB-5MS for a pesticide screen containing 100 analytes. This column produces consistent, reproducible results with very little column bleed. I recommend getting the one with the 5 meter Guardian column attached to prolong column life. Sky liners are great, I use an APEX liner which can accomodate large volume injections (up to ~250 uL I believe)
Posted: Wed Aug 01, 2012 10:47 am
by i.horsting
Mr. Brown.
Thank you for choosing OPTIC-4.
If you need any assistance, please don't hesitate to contact us.
www.atasgl.com
(i.horsting@atasgl_com) change _ to .