Concentration calculation of disodium salt standard
4 posts Page 1 of 1
Hi
I am analyzing ATP by reverse phase HPLC. I want to calculate the concentration of ATP in a sample using as reference the calibration curve of the standard (ATP disodium salt). My question is the following: Do I have to subtrac the sodium weight from the molecular weight of the standar? before to calculate the ATP concentration of the samples.
Thanks!
use a correction factor based on MW
Factor = MW ATP/MW sodium salt
weight of sodium salt X factor will give you the weight of ATP present (assuming 100% purity)
Good judgment comes from bad experience, and a lot of that comes from bad judgment.
thanks a lot!
things like that I prefer to express on a molar basis (which often makes a lot more sense anyway. Who has ever quoted the Km of an enzyme in mg/mL??). That way I can calculate the molar concentration of my standard based on whatever salt I buy, and my reader (assuming anyone reads anything I write) can't get it wrong...
4 posts Page 1 of 1
In total there are 11 users online :: 1 registered, 0 hidden and 10 guests (based on users active over the past 5 minutes)