Wide bore guard column to narrow bore true column
Posted: Wed Apr 11, 2012 6:28 pm
by ZeroAir
I am thinking about switching to a .53 guard column (larger surface area = better deposition/longer protective lifetime of/from non-volatiles), but I am using a .32 true column. I will join them with one of Agilent's deactivated glass junctions.
I sometimes have trouble sealing leaks at the junction when using .32 guard/.32 true column, and I think this problem will be exacerbated when increasing the local pressure at the junction by incorporating a .53/.32 union.
Do any of you use wider guards than your actual column and if so do you experience leaks?
Luke
Re: Wide bore guard column to narrow bore true column
Posted: Thu Apr 12, 2012 9:13 am
by Peter Apps
I have gas flow switches with three different diameters of fused silica connected with Siltek Y connectors (Restek). Once they are put together properly they do not leak.
Posted: Thu Apr 12, 2012 5:34 pm
by ZeroAir
Cool, thanks! I guess I will order the parts and give it a shot then.
Cheers,
Luke
Posted: Fri Apr 13, 2012 1:28 pm
by AICMM
zeroAir,
I have use a 0.53 to 0.25 in the past with the idea that I take more out of the inlet but still do 0.25 chromatography. Having said that, I used a glass press fit but I glued it. I also, always, recommend you do all of this with nothing hooked up so you are not fighting gas pressure while trying to make the seal.
Best regards,
AICMM
Posted: Fri Apr 13, 2012 3:15 pm
by ZeroAir
AICMM, thanks for the reply. Did you try doing that union without the glue?
Posted: Mon Apr 16, 2012 2:08 pm
by AICMM
AeroAir,
When I do press fits (not to often anymore) I always use glue. You might also consider other types such as Restek's or Gerstel's view type unions. These use ferrules to maintain pressure on the press seal.
Best regards,
AICMM