Agilent G1315 DAD Noise
Posted: Mon Apr 16, 2012 1:42 pm
I have two DAD detectors which are giving high baseline noise than they used to and it's now interferring with peak integration, causing the SSTs to fail.
I have changed the lamps (more than once) and have seen no improvment. The noise is still there even when I remove the flow cell so the rest of the HPLC system can be taken out of the equation.
I tried cleaning the windows and even exchanged the entire optical unit and saw no change. Although it didn't look like electrical interference noise I tried moving the system and that made no difference either (besides, the two troublesome units are in different rooms and there's no problem with neighbouring systems).
I haven't run an ASTM noise test as it's not running Chemstation so I haven't worked out how to do that through their Chromeleon software without running a whole PQ, however the undulations have an average height of 0.02mAU which is out of spec (should be <0.01).
Any ideas? I'm going to try taking the main board out and cleaning up the contacts but to be honest I'm not holding my breath with that solution...
Thank you.
Thank you.