PL-ELS 2100 ELSD to AGILENT 1200 (DO I NEED AN AD CONVERTER?
10 posts Page 1 of 1
Not very electronics minded but I have been thinking of buying this polymer labs PL-ELS 2100, ELSD detector to join to the existing Agilent 1200 system which currently has a DAD dector attached. Hoping to use CHemstation B.04.
I have been told that I will need to purchase an AD Converter from Agilent by one person...is this true?
SO
Yes, and an A/D license for the ChemStation B04.03 or OpenLAB CDS ChemStation, whatever you want. The ELSD is not controlled under ChemStation and therefore the signal has to be collected analog.
Freek Varossieau
OpenLab CDS 2 specialist
BeyondOpenLab
beyondopenlab@gmail.com
+5977114721
Yeah, something like a #35900E box
Thanks Guys,
Would there be an issue with Chemstation and how it handled the converted data. ie could it integrate data for example??
JC
The A/D converter type 35900E becomes an integral part of the method, as a normal channel and therefore will be treated exactly the same as all other channels. So yes you would be able to integrate this channel and assign peak names and calibration to it.
Freek Varossieau
BeyondOpenLab
+5977114721
It seems to me that The 35900E box is HPIB only, though, right? If so, I don't think you can run it from B version Chemstation (no HPIB).
I might be wrong here, I just wanted to raise this issue if it might be a problem.
Cheers,
Dwight
I Will Post The correct partnumbers for LAN based 35900
Freek Varossieau
BeyondOpenLab
+5977114721
The following part numbers you need for a LAN based 35900E A/D converter
p/n 35900E
The actual A/D converter
p/n 35900E option 010
General purpose interface cables
TCP/IP LAN MIO board for host communication
Prices to be obtained from your local Agilent representative.
Good luck
Freek Varossieau
BeyondOpenLab
+5977114721
Thanks for all your help. Will the agilent converter work with the Jet direct card...or does it need the newer ones.
JC
Cheers,
Dwight
I realise I have an 35900E and it isnt hpib - it is configured for lan
JC
Users browsing this forum: Google [Bot] and 1 guest
