Posted: Fri Jun 15, 2012 6:09 pm
Seems the lower exchange capacity column ought to offer the same N, ability to use weaker eluent for decreased background noise, and throw late-eluting anions, such as sulfate, phosphate back into the retention time window as other common ions, like Bromide and chloride.
So long as the weakest-retained anion in retained (say, F-),, why would one prefer a higher exchange capacity anion exchange column over a like column with, say, half the exchange capacity?
If there are instances were a higher exchange capacity column is more desirable, at what fold difference in capacity does this difference become non-trivial?
So long as the weakest-retained anion in retained (say, F-),, why would one prefer a higher exchange capacity anion exchange column over a like column with, say, half the exchange capacity?
If there are instances were a higher exchange capacity column is more desirable, at what fold difference in capacity does this difference become non-trivial?